CMM Manager UCC Server Setup

Post by myork1967 » Thu Dec 12, 2019 9:24 pm

Good Evening All,

Currently on a UCC Server and noticed that whoever setup the machine did not setup the table height correctly. Since the commissioning of the machine is done in the server and i cant do it in cmm manager, how would i go about (where) in server to correct this. The table appears to be about 6 inches below where it actually is.

Thanks

User avatar
CMM Guru
Moderator
Moderator
Posts: 118
Joined: Thu Feb 24, 2011 7:49 pm

Re: CMM Manager UCC Server Setup

Post by CMM Guru » Fri Jan 17, 2020 4:44 pm

Are you using I++? Or direct connect?

The machine volume is from the UCC machine.ini file (UCCassist commissioning). Oddly, the table "height" and thickness are not. They are in UCC Server and can be edited there.

Better yet? In CMM-Manager, just run the table height routine (move the probe near the table and click OK).